Biomedical Engineering > GATE 2025 > Sensors
Resistance R of a thermistor varies as a function of temperature T such that
R(T)=R0exp[β(1/T-1/T0)].
where β=3100 K and, R0 and T0 are positive constants. If the relative error in measuring R is 10%, what is the relative error (in percentage) in measuring 310 K temperature? (rounded off to the nearest integer)
R(T)=R0exp[β(1/T-1/T0)].
where β=3100 K and, R0 and T0 are positive constants. If the relative error in measuring R is 10%, what is the relative error (in percentage) in measuring 310 K temperature? (rounded off to the nearest integer)
Correct : 3