Biomedical Engineering > GATE 2024 > EEG
A neurologist needs to observe the alpha wave in EEG recordings of a patient. The system block diagram with ideal filter blocks is shown below. Which one of the following design choices is correct?
A
fh=8 Hz fl=12 Hz fs=12 Hz
B
fh=4 Hz. fl=6 Hz fs=24 Hz
C
fh=6 Hz, fl=4Hz, fs=12 Hz
D
fh=8 Hz., fl=12 Hz, fs=48 Hz

Correct : d