A Wheatstone bridge strain gauge transducer is constructed on a diaphragm in such a way that when a force is applied on the diaphragm, the resistors R1 and R4 will be in compression, and the resistors R2 and R3 will be in tension. The bridge excitation voltage (Ein) is 10 Volts. If all the resistors have a resistance of 200 Ω in the absence of any force, and each resistance changes by 20 Ω upon application of a force, what is the output voltage Vout (in Volts) from the Wheatstone bridge? (Round off your answer to the nearest integer.)

Correct : 1