Biomedical Engineering > GATE 2023 > Electrocardiography
At time t, the cardiac dipole is oriented at -45° (minus forty five degrees) to the horizontal axis. The magnitude of the dipole is 3 mV. Assuming Einthoven frontal plane configuration, what is the magnitude (in mV) of the electrical signal in lead II? (Round off the answer to two decimal places.)
Correct : 3.18