Biomedical Engineering > GATE 2020 > Blood Flow Measurement
In the process of measuring blood flow from an artery using C-clamp magnetic flow probe, the voltage recorded across diametrically opposite sites of the artery is 3.75 nV. The blood flow rate through the artery is _______ cm³/s (rounded off to two decimal places).
The inner diameter of the C-clamp = 0.5 cm,
The magnetic flux density = 1.5 × 10⁻⁵ Wb/m²

Correct : 0.95 to 0.99