Biomedical Engineering > GATE 2020 > Cardiac Pacing
A 100 mA single current pulse from a pulse generator is used for artificial pacing of heart at the right ventricle. If the delivered energy does not exceed the fibrillation threshold of 300 μJ, the safe duration of the pulse that could be applied to the tissue mass having an impedance of 500 Ω is _______ μs.

Correct : 60 to 60